Element 3 Question 3-75L4
C3L75: Test Equipment
On an analog wattmeter, what part of the scale is most accurate and how much does that accuracy extend to the rest of the reading scale?
- The accuracy is only at full scale, and that absolute number reading is carried through to the rest of the range. The upper 1/3 of the meter is the only truly calibrated part. (correct answer)
- The accuracy is constant throughout the entire range of the meter.
- The accuracy is only there at the upper 5% of the meter, and is not carried through at any other reading.
- The accuracy cannot be determined at any reading.
Correct answer: A, The accuracy is only at full scale, and that absolute number reading is carried through to the rest of the range. The upper 1/3 of the meter is the only truly calibrated part.
Explanation
Analog meter accuracy is stated as a percentage of full scale. A ±5% rating on a 100 W scale means ±5 W anywhere on the scale. That fixed error is small at 90 W but large at 10 W, where it is 50% of the reading. Readings are trustworthy mainly in the upper third of the scale, so choose a range that puts the reading there.
Why the other answers are wrong
- B, The accuracy is constant throughout the entire range of the meter.: The absolute error stays constant, but the error as a share of the reading grows toward the low end. Accuracy is not constant across the scale.
- C, The accuracy is only there at the upper 5% of the meter, and is not carried through at any other reading.: The full-scale accuracy figure does carry to other readings as a fixed absolute error. Accuracy is not limited to the top 5%.
- D, The accuracy cannot be determined at any reading.: The accuracy can be determined from the full-scale rating. It just becomes poorer, relative to the reading, at low deflections.