C3L75: Test Equipment

On an analog wattmeter, what part of the scale is most accurate and how much does that accuracy extend to the rest of the reading scale?

  1. The accuracy is only at full scale, and that absolute number reading is carried through to the rest of the range. The upper 1/3 of the meter is the only truly calibrated part. (correct answer)
  2. The accuracy is constant throughout the entire range of the meter.
  3. The accuracy is only there at the upper 5% of the meter, and is not carried through at any other reading.
  4. The accuracy cannot be determined at any reading.

Correct answer: A, The accuracy is only at full scale, and that absolute number reading is carried through to the rest of the range. The upper 1/3 of the meter is the only truly calibrated part.

Explanation

Analog meter accuracy is stated as a percentage of full scale. A ±5% rating on a 100 W scale means ±5 W anywhere on the scale. That fixed error is small at 90 W but large at 10 W, where it is 50% of the reading. Readings are trustworthy mainly in the upper third of the scale, so choose a range that puts the reading there.

Why the other answers are wrong