Amateur Extra Question E4A04

E4A: Test equipment: analog and digital instruments; spectrum analyzers; antenna analyzers; oscilloscopes; RF measurements

How is compensation of an oscilloscope probe performed?

  1. A square wave is displayed, and the probe is adjusted until the horizontal portions of the displayed wave are as nearly flat as possible (correct answer)
  2. A high frequency sine wave is displayed, and the probe is adjusted for maximum amplitude
  3. A frequency standard is displayed, and the probe is adjusted until the deflection time is accurate
  4. A DC voltage standard is displayed, and the probe is adjusted until the displayed voltage is accurate

Correct answer: A, A square wave is displayed, and the probe is adjusted until the horizontal portions of the displayed wave are as nearly flat as possible

Explanation

A 10X probe has an adjustable capacitor that balances its divider against the scope's input capacitance. Connect the probe to a square wave, usually the scope's built-in calibrator. Adjust until the tops and bottoms are flat, with no overshoot (overcompensated) and no rounded corners (undercompensated). A square wave works well because it contains many harmonics, so any error in high-frequency response is easy to see.

Why the other answers are wrong