Which of the following is good practice when using an oscilloscope probe?
- Minimize the length of the probe's ground connection (correct answer)
- Never use a high-impedance probe to measure a low-impedance circuit
- Never use a DC-coupled probe to measure an AC circuit
- All these choices are correct
Correct answer: A, Minimize the length of the probe's ground connection
Explanation
The probe's ground lead is a small inductor, and together with the probe's input capacitance it forms a resonant circuit. A long ground lead adds inductance and picks up stray fields, so fast edges show ringing, overshoot and noise that aren't really in the circuit. Keeping the ground connection as short as possible, such as with a short spring tip clipped right at the test point, keeps the displayed waveform faithful.
Why the other answers are wrong
- B, Never use a high-impedance probe to measure a low-impedance circuit: High-impedance probes are standard for low-impedance circuits because they barely load the circuit, so there is no rule against using one there.
- C, Never use a DC-coupled probe to measure an AC circuit: DC coupling passes AC signals as well as any DC offset; you choose AC coupling only to remove the DC part, not because DC coupling can't display AC.
- D, All these choices are correct: Only the short-ground-lead advice is valid; the statements about high-impedance probes and DC coupling are both false.