Amateur Extra Question E4B05
E4B: Measurement technique and limitations: instrument accuracy and performance limitations; probes; techniques to minimize errors; measurement of Q; instrument calibration; S parameters; vector network analyzers; RF signals
What three test loads are used to calibrate an RF vector network analyzer?
- 50 ohms, 75 ohms, and 90 ohms
- Short circuit, open circuit, and 50 ohms (correct answer)
- Short circuit, open circuit, and resonant circuit
- 50 ohms through 1/8 wavelength, 1/4 wavelength, and 1/2 wavelength of coaxial cable
Correct answer: B, Short circuit, open circuit, and 50 ohms
Explanation
A VNA is calibrated with short, open and load (SOL) standards. The load matches the system impedance, normally 50 Ω. The short reflects everything with one phase, the open reflects everything with the opposite phase, and the load reflects nothing. These three known references let the analyzer solve for and remove its own errors and those of the test cable, which places the measurement reference plane at the connector.
Why the other answers are wrong
- A, 50 ohms, 75 ohms, and 90 ohms: Three resistive loads near 50 Ω don't give the widely separated, known reflection points needed to solve the error model.
- C, Short circuit, open circuit, and resonant circuit: A resonant circuit's impedance changes sharply with frequency, so it can't serve as a known reference across a sweep.
- D, 50 ohms through 1/8 wavelength, 1/4 wavelength, and 1/2 wavelength of coaxial cable: Line sections act as standards only in special methods; the standard short-open-load calibration doesn't use 1/8, 1/4 and 1/2 wavelength cables.